SAMPLE PREPARATION OF BLOOD TO ANALYSE HEAVY METAS USING ATOMIC ABSORPTION SPECTRSCOPY

2006-06-27 04:38:20 · 2 answers · asked by KRISH 1 in Science & Mathematics Chemistry

2 answers

I work in a petroleum testing laboratory and we use an AAS on a daily basis. Although we don't test blood, I would assume you would either dilute a known weighted sample of blood in a blank solution (in my case we use a Kerosene/Xylene solution) and aspirate as normal and make our calculations from the absorbances of 3 standards and our test sample. The other way we run a sample on our AAS is to ash the sample. We burn it in a 750 degree celcius oven for approx 4 hours, after cooling, flux it with a 90/10 mixture of powdered chemicals (the names I cannot remember) in a 950 degree celcius oven for 10 minutes, boil the ash away with tartaric acid solution for 5 minutes, then make remaining solution aqueous by adding DI water up to a 100 mL flask. We would run that sample against 3 known standards and make our calculation from that.

2006-06-27 04:59:45 · answer #1 · answered by Anonymous · 0 0

No special preparation required.

2006-06-27 04:40:19 · answer #2 · answered by ag_iitkgp 7 · 0 0